iV/X10 Setup Suggestions

Users Who Are Viewing This Thread (Total: 1, Members: 0, Guests: 1)

JeffB418

HYDROS Expert
Joined
Jun 25, 2020
Messages
1,157
Reaction score
337
Location
Syracuse NY
Here are a few suggestions for those running the iV/X10 to improve performance based on my personal experience with the X10:

Having accuracy issues?
- Read the suggestions below.
- Make sure your dosing tubes are fully placed onto the dosing head nipples as far as possible! Prevents air leaks and issues.
- Make sure your schedule is testing atleast once a day, I recommend minimum of 4 tests a day.
- Run the "unkink tubing" from the pump status pop up on all iV pumps a few times to work new/ridged pump heads. This will help work brand new heads free after shipping or replacement.


iV Placement:
- For BEST performance, place the iV close to the X10 and the dosing pumps. This will help improve the accuracy of the dosers. The most important portion of the setup are the dosing lines from your reagent vessel, to the X10 and to the iV. Keeping these as short as possible will improve precision.
- I personally recommend placing the iV right next to the X10.
- Also keep the sample line as short as possible, but under 3ft.

Doser Calibration Recommendations:
- USE A SCALE! Biggest thing that can help you get the most accurate results is to use a precision scale and weigh your calibration fluids. I suggest using something like this scale wehre it has a 0.1G resolution: Amazon.com
- When using the scale, for reagent calibration use 1 gram = 1 mL, since the density of reagent is very close to the density of water.
- When using the scale, for calibrating the fill and drain dosers using tank water, you have to know the SG of your salt water to figure out the gram to ml conversion. If your saltwater is 1.025, then take the weight of your calibration sample and divide it by your salinity. IE: (weight of your sample in grams)/(SG of your tank)= volume in mL. Otherwise if you use gram = ml, your calibration will be off!
- Calibration of your reagent is the most important, followed by calibration of the fill doser. Calibration of the drain doser doesn't have to be perfect since it just drains, but it has to be in the ball park so it knows it actually clears the beaker fully.
- When setting up for the first time, you will need to recalibrate your dosers after a few days since the pumps/tubing need to break in. Then make sure to calibrate monthy after to maintain accuracy.

pH Probe Recommendations:
- The pinpoint probe works with the iV but I personally suggest using a glass probe like this model below. Seems to be more responsive and less affected by testing reagent:
Milwaukee MA991B/1 Glass pH Replacement Probe
- Calibrate using pH 4 and 7 solutions, 10 is optional.
- Calibrate after a week of initial setup since some probes need to break in

Doser Head Recommendations:
- Replace your X10 dosing heads every so often since they do wear down, depending on how often you test, but 6 months is a good interval.

Regent Mixing:
- ONLY mix reagent that says its "concentrated". Ready to use reagent does not need to be diluted, and concentrated reagent MUST be diluted.

- When making your own reagent from the concentrated version (Alkalinity Testing Machine Concentrated Reagent) precision in your measuring equipment is KEY!
- The MOST accurate method of dilution, is using a volumetric flask. These are available for cheap on amazon. Use a volume that is 1/5th of your target volume size. So if you want to mix 1 gallon of reagent, a 500mL volumetric flask. And use the SAME flask for the concentrate as the RODI. This will produce a reagent volume of 2500mL or 0.7 Gallons.
- Remember to dilute using 1 part concentrated reagent to 4 parts RODI water. When using your volumetric flask, you will fill it once with the concentrated reagent and then 4 times with RODI.

Reagent Testing:
- There is a way to use Salifert Alk testing kit to verify your diluted reagent.
- First test your sample water using the salifert kit as directed, note the ALK value produced by the test.
- Second, re-test again using fresh sample water but this time instead of using the salifert liquid reagent, use your mixed iV reagent in the syringe. Run the test the same as before, ended at the same color conversion. Then note how much reagent from the syringe was USED, IE if the syringe has 0.4mL left, you USED 0.6ml. Take the USED value and calculate dKH using this equation: dKH = iV Reagent Used (mL) x 250 x 0.056.
- Compare the ALK value generated by the standard salifert test and the second test with iV reagent. If your reagent is mixed correctly, these should be close. If your iV test dKH is really high or low compared to the salifert baseline, your reagent is mixed wrong.


Finally make sure to completely read the setup guide here: HYDROS Control - How To Add an iV Tester Output
 
Last edited:
Is getting the Milwaukee pH probe necessary? I just feel a bit cheated since I set up everything for a client and now I am realizing that a different probe is recommended instead of the one provided... I'm more so wondering about the consistency of pH and alk results. If it is a bit skewed that's fine as long as it is consistent.
 
Is getting the Milwaukee pH probe necessary? I just feel a bit cheated since I set up everything for a client and now I am realizing that a different probe is recommended instead of the one provided... I'm more so wondering about the consistency of pH and alk results. If it is a bit skewed that's fine as long as it is consistent.
I use the Pinpoint probes with the X10/iV and I have had no issues with it.
Here is a graph of the pH probe in the sump and the one in the iV. The iV one takes a sample at the start of the alkalinity test which is done on mine 4 times a day. It seems to track close to the one in the sump at the time of the sample.

IMG_9917.png
 
I use the Pinpoint probes with the X10/iV and I have had no issues with it.
Here is a graph of the pH probe in the sump and the one in the iV. The iV one takes a sample at the start of the alkalinity test which is done on mine 4 times a day. It seems to track close to the one in the sump at the time of the sample.

View attachment 3915
Thanks so much. Yeah, then I will just stick to the current probe and maybe swap it out in the future if needed.
 
Here are a few suggestions for those running the iV/X10 to improve performance based on my personal experience with the X10:

Having accuracy issues?
- Read the suggestions below.
- Make sure your dosing tubes are fully placed onto the dosing head nipples as far as possible! Prevents air leaks and issues.
- Make sure your schedule is testing atleast once a day, I recommend minimum of 4 tests a day.
- Run the "unkink tubing" from the pump status pop up on all iV pumps a few times to work new/ridged pump heads. This will help work brand new heads free after shipping or replacement.


iV Placement:
- For BEST performance, place the iV close to the X10 and the dosing pumps. This will help improve the accuracy of the dosers. The most important portion of the setup are the dosing lines from your reagent vessel, to the X10 and to the iV. Keeping these as short as possible will improve precision.
- I personally recommend placing the iV right next to the X10.
- Also keep the sample line as short as possible, but under 3ft.

Doser Calibration Recommendations:
- USE A SCALE! Biggest thing that can help you get the most accurate results is to use a precision scale and weigh your calibration fluids. I suggest using something like this scale wehre it has a 0.1G resolution: Amazon.com
- When using the scale, for reagent calibration use 1 gram = 1 mL, since the density of reagent is very close to the density of water.
- When using the scale, for calibrating the fill and drain dosers, you have to know the SG of your salt water to figure out the gram to ml conversion. If your saltwater is 1.025, then take the weight of your calibration sample and divide it by your salinity. IE: (weight of your sample in grams)/(SG of your tank)= volume in mL. Otherwise if you use gram = ml, your calibration will be off!
- Calibration of your reagent is the most important, followed by calibration of the fill doser. Calibration of the drain doser doesn't have to be perfect since it just drains, but it has to be in the ball park so it knows it actually clears the beaker fully.
- When setting up for the first time, you will need to recalibrate your dosers after a few days since the pumps/tubing need to break in. Then make sure to calibrate monthy after to maintain accuracy.

pH Probe Recommendations:
- The pinpoint probe works with the iV but I personally suggest using a glass probe like this model below. Seems to be more responsive and less affected by testing reagent:
Milwaukee MA991B/1 Glass pH Replacement Probe
- Calibrate using pH 4 and 7 solutions, 10 is optional.
- Calibrate after a week of initial setup since some probes need to break in

Doser Head Recommendations:
- Replace your X10 dosing heads every so often since they do wear down, depending on how often you test, but 6 months is a good interval.

Regent Mixing:
- ONLY mix reagent that says its "concentrated". Ready to use reagent does not need to be diluted, and concentrated reagent MUST be diluted.

- When making your own reagent from the concentrated version (Alkalinity Testing Machine Concentrated Reagent) precision in your measuring equipment is KEY!
- The MOST accurate method of dilution, is using a volumetric flask. These are available for cheap on amazon. Use a volume that is 1/5th of your target volume size. So if you want to mix 1 gallon of reagent, a 500mL volumetric flask. And use the SAME flask for the concentrate as the RODI. This will produce a reagent volume of 2500mL or 0.7 Gallons.
- Remember to dilute using 1 part concentrated reagent to 4 parts RODI water. When using your volumetric flask, you will fill it once with the concentrated reagent and then 4 times with RODI.

Reagent Testing:
- There is a way to use Salifert Alk testing kit to verify your diluted reagent.
- First test your sample water using the salifert kit as directed, note the ALK value produced by the test.
- Second, re-test again using fresh sample water but this time instead of using the salifert liquid reagent, use your mixed iV reagent in the syringe. Run the test the same as before, ended at the same color conversion. Then note how much reagent from the syringe was USED, IE if the syringe has 0.4mL left, you USED 0.6ml. Take the USED value and calculate dKH using this equation: dKH = iV Reagent Used (mL) x 250 x 0.056.
- Compare the ALK value generated by the standard salifert test and the second test with iV reagent. If your reagent is mixed correctly, these should be close. If your iV test dKH is really high or low compared to the salifert baseline, your reagent is mixed wrong.


Finally make sure to completely read the setup guide here: HYDROS Control - How To Add an iV Tester Output
When calibrating the fill and drain,
Is that calibrating with rodi water and then taking the weighed sample and dividing by the salinity?
Or is that calibrating with tank water , and taking that weighed sample and dividing it by the salinity and that number is what gets typed in on the calibration page ?

I want to make sure and understand I’m doing this correctly as I’m getting various different alk readings everytime it test .
 
When calibrating the fill and drain,
Is that calibrating with rodi water and then taking the weighed sample and dividing by the salinity?
Or is that calibrating with tank water , and taking that weighed sample and dividing it by the salinity and that number is what gets typed in on the calibration page ?

I want to make sure and understand I’m doing this correctly as I’m getting various different alk readings everytime it test .
always calibrate with tank water since thats what the iV uses during tests. and yes, take the weighed sample and divide it and then use that result in the app.
 
I’d like to use the 4th available pump to dose alk… what’s the lowest flow rate these go?
 
Here are a few suggestions for those running the iV/X10 to improve performance based on my personal experience with the X10:

Having accuracy issues?
- Read the suggestions below.
- Make sure your dosing tubes are fully placed onto the dosing head nipples as far as possible! Prevents air leaks and issues.
- Make sure your schedule is testing atleast once a day, I recommend minimum of 4 tests a day.
- Run the "unkink tubing" from the pump status pop up on all iV pumps a few times to work new/ridged pump heads. This will help work brand new heads free after shipping or replacement.


iV Placement:
- For BEST performance, place the iV close to the X10 and the dosing pumps. This will help improve the accuracy of the dosers. The most important portion of the setup are the dosing lines from your reagent vessel, to the X10 and to the iV. Keeping these as short as possible will improve precision.
- I personally recommend placing the iV right next to the X10.
- Also keep the sample line as short as possible, but under 3ft.

Doser Calibration Recommendations:
- USE A SCALE! Biggest thing that can help you get the most accurate results is to use a precision scale and weigh your calibration fluids. I suggest using something like this scale wehre it has a 0.1G resolution: Amazon.com
- When using the scale, for reagent calibration use 1 gram = 1 mL, since the density of reagent is very close to the density of water.
- When using the scale, for calibrating the fill and drain dosers using tank water, you have to know the SG of your salt water to figure out the gram to ml conversion. If your saltwater is 1.025, then take the weight of your calibration sample and divide it by your salinity. IE: (weight of your sample in grams)/(SG of your tank)= volume in mL. Otherwise if you use gram = ml, your calibration will be off!
- Calibration of your reagent is the most important, followed by calibration of the fill doser. Calibration of the drain doser doesn't have to be perfect since it just drains, but it has to be in the ball park so it knows it actually clears the beaker fully.
- When setting up for the first time, you will need to recalibrate your dosers after a few days since the pumps/tubing need to break in. Then make sure to calibrate monthy after to maintain accuracy.

pH Probe Recommendations:
- The pinpoint probe works with the iV but I personally suggest using a glass probe like this model below. Seems to be more responsive and less affected by testing reagent:
Milwaukee MA991B/1 Glass pH Replacement Probe
- Calibrate using pH 4 and 7 solutions, 10 is optional.
- Calibrate after a week of initial setup since some probes need to break in

Doser Head Recommendations:
- Replace your X10 dosing heads every so often since they do wear down, depending on how often you test, but 6 months is a good interval.

Regent Mixing:
- ONLY mix reagent that says its "concentrated". Ready to use reagent does not need to be diluted, and concentrated reagent MUST be diluted.

- When making your own reagent from the concentrated version (Alkalinity Testing Machine Concentrated Reagent) precision in your measuring equipment is KEY!
- The MOST accurate method of dilution, is using a volumetric flask. These are available for cheap on amazon. Use a volume that is 1/5th of your target volume size. So if you want to mix 1 gallon of reagent, a 500mL volumetric flask. And use the SAME flask for the concentrate as the RODI. This will produce a reagent volume of 2500mL or 0.7 Gallons.
- Remember to dilute using 1 part concentrated reagent to 4 parts RODI water. When using your volumetric flask, you will fill it once with the concentrated reagent and then 4 times with RODI.

Reagent Testing:
- There is a way to use Salifert Alk testing kit to verify your diluted reagent.
- First test your sample water using the salifert kit as directed, note the ALK value produced by the test.
- Second, re-test again using fresh sample water but this time instead of using the salifert liquid reagent, use your mixed iV reagent in the syringe. Run the test the same as before, ended at the same color conversion. Then note how much reagent from the syringe was USED, IE if the syringe has 0.4mL left, you USED 0.6ml. Take the USED value and calculate dKH using this equation: dKH = iV Reagent Used (mL) x 250 x 0.056.
- Compare the ALK value generated by the standard salifert test and the second test with iV reagent. If your reagent is mixed correctly, these should be close. If your iV test dKH is really high or low compared to the salifert baseline, your reagent is mixed wrong.


Finally make sure to completely read the setup guide here: HYDROS Control - How To Add an iV Tester Output
How do you feel about dilution by weight for the reagent concentrate?
 
Also, any issue with material compatibility for storing the diluted reagent? Distilled water 1 gallon jug/clean milk jug from the store fine?
 
Also, any issue with material compatibility for storing the diluted reagent? Distilled water 1 gallon jug/clean milk jug from the store fine?
should be ok. as long as they are distilled water jugs and not milk jugs. I personally prefer HDPE plastic jugs from amazon
 
Here are a few suggestions for those running the iV/X10 to improve performance based on my personal experience with the X10:

Having accuracy issues?
- Read the suggestions below.
- Make sure your dosing tubes are fully placed onto the dosing head nipples as far as possible! Prevents air leaks and issues.
- Make sure your schedule is testing atleast once a day, I recommend minimum of 4 tests a day.
- Run the "unkink tubing" from the pump status pop up on all iV pumps a few times to work new/ridged pump heads. This will help work brand new heads free after shipping or replacement.


iV Placement:
- For BEST performance, place the iV close to the X10 and the dosing pumps. This will help improve the accuracy of the dosers. The most important portion of the setup are the dosing lines from your reagent vessel, to the X10 and to the iV. Keeping these as short as possible will improve precision.
- I personally recommend placing the iV right next to the X10.
- Also keep the sample line as short as possible, but under 3ft.

Doser Calibration Recommendations:
- USE A SCALE! Biggest thing that can help you get the most accurate results is to use a precision scale and weigh your calibration fluids. I suggest using something like this scale wehre it has a 0.1G resolution: Amazon.com
- When using the scale, for reagent calibration use 1 gram = 1 mL, since the density of reagent is very close to the density of water.
- When using the scale, for calibrating the fill and drain dosers using tank water, you have to know the SG of your salt water to figure out the gram to ml conversion. If your saltwater is 1.025, then take the weight of your calibration sample and divide it by your salinity. IE: (weight of your sample in grams)/(SG of your tank)= volume in mL. Otherwise if you use gram = ml, your calibration will be off!
- Calibration of your reagent is the most important, followed by calibration of the fill doser. Calibration of the drain doser doesn't have to be perfect since it just drains, but it has to be in the ball park so it knows it actually clears the beaker fully.
- When setting up for the first time, you will need to recalibrate your dosers after a few days since the pumps/tubing need to break in. Then make sure to calibrate monthy after to maintain accuracy.

pH Probe Recommendations:
- The pinpoint probe works with the iV but I personally suggest using a glass probe like this model below. Seems to be more responsive and less affected by testing reagent:
Milwaukee MA991B/1 Glass pH Replacement Probe
- Calibrate using pH 4 and 7 solutions, 10 is optional.
- Calibrate after a week of initial setup since some probes need to break in

Doser Head Recommendations:
- Replace your X10 dosing heads every so often since they do wear down, depending on how often you test, but 6 months is a good interval.

Regent Mixing:
- ONLY mix reagent that says its "concentrated". Ready to use reagent does not need to be diluted, and concentrated reagent MUST be diluted.

- When making your own reagent from the concentrated version (Alkalinity Testing Machine Concentrated Reagent) precision in your measuring equipment is KEY!
- The MOST accurate method of dilution, is using a volumetric flask. These are available for cheap on amazon. Use a volume that is 1/5th of your target volume size. So if you want to mix 1 gallon of reagent, a 500mL volumetric flask. And use the SAME flask for the concentrate as the RODI. This will produce a reagent volume of 2500mL or 0.7 Gallons.
- Remember to dilute using 1 part concentrated reagent to 4 parts RODI water. When using your volumetric flask, you will fill it once with the concentrated reagent and then 4 times with RODI.

Reagent Testing:
- There is a way to use Salifert Alk testing kit to verify your diluted reagent.
- First test your sample water using the salifert kit as directed, note the ALK value produced by the test.
- Second, re-test again using fresh sample water but this time instead of using the salifert liquid reagent, use your mixed iV reagent in the syringe. Run the test the same as before, ended at the same color conversion. Then note how much reagent from the syringe was USED, IE if the syringe has 0.4mL left, you USED 0.6ml. Take the USED value and calculate dKH using this equation: dKH = iV Reagent Used (mL) x 250 x 0.056.
- Compare the ALK value generated by the standard salifert test and the second test with iV reagent. If your reagent is mixed correctly, these should be close. If your iV test dKH is really high or low compared to the salifert baseline, your reagent is mixed wrong.


Finally make sure to completely read the setup guide here: HYDROS Control - How To Add an iV Tester Output
Jeff are you still happy with the Milwaukee glass probe as a recommended replacement for the Pinpoint?
 