Danny
HYDROS Expert
That is quite a difference. How many test have been done since the tubing issue was corrected? There may have been some air bubbles in the line. That will cause high readings like that. ai would give it at least three test before ruling that out. Even better would be to watch to see if you see any air bubbles in the line. Any air bubble will cause it to dispense less reagent than it has calculated and cause a higher reading than it actually is. If that is not the issue them run a manual dispense of 3ml in a beaker where it can be weighed to see if you are actually dispensing 3ml. If it is less than that then do a calibration. I doubt it would be more with that high a reading. After that recheck the pH calibration.